Dear All,
We have compiled a list of the most common fixes on the DeltaVisions
when "the fluorescence is weak" or "I can only see part of the field
of view" or "the green and the red should be in the same place and
they're not". This is a list of simple things you can check yourself
but, as ever, please come and ask us if you want any help
whatsoever. If any of you have an interest in what these bits
actually do, please pop down for a chat. We even have a coffee machine.
Could I also again remind people doing DIC and experiments with the
QLM to put the microscope back to the "standard configuration" after
use i.e. prism and analyser out, laser mirror out.
Cheers!
Sam
If you are NOT doing Kohler, DIC or a Photokintetics experiment, and
you have problems, check the following:
Clean the lens (an oil bubble will cause weak fluorescence or an
obstructed field of view)
Fluorescence Shutter under objective lens needs to be OPEN (otherwise
you won't see any fluorescence)
Dichroic mirror in correct position and locked in place (will cause a
weak signal, no signal, or an obstructed field of view)
DIC prism and analyser should be OUT (or you'll lose lots of trans or
fluorescence light, plus have all kinds of weird stuff going on in
fluorescence)
Laser mirror should be in OUT position (or you'll cut some light out
in fluorescence)
UV blocking filter should be in OUT position (or you won't see DAPI
or much CFP)
Transmitted light aperture needs to be OPEN (if the condenser is out
of alignment and this is closed, you won't see any transmitted light)
The eyepiece lens should be set to 0 (ie the Bertrand lens – CT
position – should not be in place, otherwise you'll be looking at the
back of the lens, not the top where your sample is)
Aux Mag should be OUT (i.e. a 1x mag NOT 1.5x mag, you'll lose light
and your sample will look very big, the stage won't move properly etc)
If you still have a problem with “alignment” or “some sort of circle
eclipsing your image” check the lens, eyepiece filter and photoframe
are properly seated (is the lens locked in place, is the photoframe
halfway out, is the eyepiece filter locked in place?)
If you have done Kohler, DIC or a Photokinetics experiment, you MUST
leave the system in the above configuration when you have finished.